WHERE CAN I FIND THE CHEMISTRY INFORMATION ABOUT GC DERIVATIVES? I.E. THE STRUCTURE OF PFPA, HFBA, MSTFA, BSTFA, ETC AND THIERE REACTIONS WITH COULMS STATIONARY PHASE
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Search vendors of GCs and associated chemicals. There are plenty of applications and descriptions of their chemicals (may have to look at MSDSs for structures, formula weights).
Agilent,
Phenomenex,
Sigma/Aldrich,
Pierce etc. etc. etc.
Agilent,
Phenomenex,
Sigma/Aldrich,
Pierce etc. etc. etc.
Thanks,
DR

DR

-
- Posts: 247
- Joined: Sat Apr 28, 2007 3:24 pm
There is a lot of information here:
http://www.sigmaaldrich.com/etc/mediali ... p/4537.pdf
As for reactions with the column stationary phase, you probably will not find that information online. There are a couple of books on derivatization reagents and reactions that would have more info, check Amazon.
http://www.sigmaaldrich.com/etc/mediali ... p/4537.pdf
As for reactions with the column stationary phase, you probably will not find that information online. There are a couple of books on derivatization reagents and reactions that would have more info, check Amazon.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
You are thinking of interactions with the stat. phase? You wouldn´t want reactions.
There are books on derivatization, I don´t have a ref. right here, but these should be easy to find, if I remember correctly there are some refs in this forum.
There are books on derivatization, I don´t have a ref. right here, but these should be easy to find, if I remember correctly there are some refs in this forum.
-
- Posts: 10
- Joined: Tue Jan 06, 2009 5:43 pm
In general if you use derivatization methods that produce nonpolar derivatives, you will have good peak shapes and less interactions with nonpolar column phases. For example, if you prepare trimethylsilyl derivatives, phases such as DB-1 or DB-5 tend to work well. Silyl ethers or silyl esters sometimes do not have good peak shapes on some polar columns such as DB-Wax (polyethylene glycol type).
When analyzing silyl ether or ester derivatives, it is also to good to minimize active sites on the column, etc., as much as possible. Things such as connectors between retention gaps and columns can lead to active silica being exposed which can cause the derivatives to degrade.
When analyzing silyl ether or ester derivatives, it is also to good to minimize active sites on the column, etc., as much as possible. Things such as connectors between retention gaps and columns can lead to active silica being exposed which can cause the derivatives to degrade.
5 posts
Page 1 of 1
Who is online
In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 27 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 27 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science