Predicting Backpressure from columns

Posted: Sun Feb 01, 2009 5:00 pm
by Pierre
Hi All,

I am using Agilent6410QQQ and want to switch methods from 4.6x150mm 5micron column to 2.1x150 1.8micron. I am however concern about the backpressure and I want to know if there is some equation that will let me know what pressure to expect.

Thanks in avance

(P.S. I am new to the forum so please be nice with the responses :lol: )

Posted: Sun Feb 01, 2009 5:33 pm
by zokitano
Hi Pierre

Welcome onboard :D

I found this on the Agilent site: RRLC Online Method Translator

http://www.chem.agilent.com/en-US/produ ... 60931.aspx

Click on the Advanced mode and there you can predict your system pressure with the new column

Hope this helps

Posted: Sun Feb 01, 2009 6:05 pm
by Pierre
Hi Pierre

Welcome onboard :D

I found this on the Agilent site: RRLC Online Method Translator

http://www.chem.agilent.com/en-US/produ ... 60931.aspx

Click on the Advanced mode and there you can predict your system pressure with the new column

Hope this helps
Thanks alot Zokitano.

Posted: Wed Feb 04, 2009 2:39 pm
by Bryan Evans
Pierre -

Are you trying to get better separation or decrease run time?
What is the optimal flow rate of your MS detector?
What is the max pressure of your LC?

(why does the link say to use 0.21mL/min for 1.8um column?)