AA » Wed Feb 04, 2009 2:12 pm
What I said was for a 1 sec wide peak, 40 Hz is high enough. Also peak width is primarily influenced on what is happening prior to the detector, system volume, flow rate, column i.d., column packing diameter,post column dispersion, chromatographic behavior of the compounds, and other stuff too. Certainly, detectors (any detector) can, and do add some peak width physically and electronicly (data rate and filter constant can appear to add very large amounts of peak width if chosen poorly).
Why do instrument vendors have higher than needed data rates? 1: Planning for the future, a detector lifespan should be many years, who knows how low peak widths will go, if you manage to generate a 0.25 sec peak, then you will need 80 Hz or higher. 2:Competitive advantage, some customers sometimes think in very simple terms, i.e. faster is better, so having the system with the highest data rate can sometimes give a slight competitive advanatge.
And yes there are compromises at higher data rates, mainly higher baseline noise.