We are looking at acetone and DMF in a pharmaceutical API. Solvent is DMA and internal std is tert-butanol. We seem to have a strange carry over problem with the DMF.
We have PE headspace and Agilent 6890 GC running Empower CDM as instrument control and data acquisition.
The DMF does not appear to be a solvent contamination.
Is it ok to run a split ratio of 0.1 to 1 or might this lead to carry-over?
We do not have any problem with acetone of tert-butanol. Peak area for these two are very consistent but the DMF response is all over the place (RSD of 6 injection around 11%). Running a 1/10 dilution of the Std gives us 1/10 the area of the acetone peak but ½ the DMF peak (with RSD of 6 = 30%). Decreasing area for the diluted std kind of point towards carry-over, but solvent blank injections between working stds and dilution do not show DMF peak.
HS conditions:
Sample oven: 120C
Needle: 165C
Transfer line: 165C
Cycle time: 30min
Thermostatting: 10min
Pressurisation: 1min
Injection: 0.1min
Withdrawal: 0.2min
Gas pressure: 15psi
Vial venting: on
HS mode: constant
Sample Shaker: on.
GC Conditions:
Column: DB-624, 30mx0.53mm, 3um film
Carrier: Helium @ 4mL/min
Oven: 40C for 5min
Ramp: 20C/min to 150C, hold 3min, 30C/min to 220C, hold 5min
Inlet temp: 165C
Detector (FID): 250C