krickos » Mon Jan 12, 2009 12:00 pm
Hi zlb215
Well personally I am quite reluctant to that as a routine approach especially if both raw data and "modified" raw data is not easily availble and gone throught during review (basic CFR/EU GMP requirement).
To start from the top so to speak. ICH Q2 R1 (analytical validation), essentially points to the pharmacopieas when it comes to system suitability test, so essentially for all releated substances methods signal to noise testing is mandatory (USP/Ph Eur) regardless if it is a generic or new product/API. Can not see that you can use modified data to comply with signal to noise testing to start with in my opinion. But of course SSTs can be done the normal way.
On the other hand, if validation/development has failed to sufficiently remove interfering peaks/baseline disturbances that are repeateble in blank runs throughout the analytical sequence, I can see the the use of using an "automated" function to remove those peaks before integration rather than manually recalculate peaks for each chromatogram.
However this would require some very clear routines around working with this and injections of blanks throughout the sequence. Also (as I do assists in audits sometimes), I would expect the unmodified raw data to be easily accesible during an audit to be able to review/verify that the SOPs/methods are followed accordingly as this is fundementaly different from disregarding peaks under the report/disregard limit.
Looking at the regulatory requirements: Regardless if you look at EU (Eudralex Volume 4, part II APIs §6.60) or CFR§211.194, a complete record of data generated should be included, including calculations/conversion factors.
I would say the situation you describe falls within this and you have to be able to show how you have done it for each and every batch.
Intresting question, I hope more provide feedback