The compound is basic in nature and its molecular weight is about 500. The aim of the method is estimate the compound as such and to separate the impurity (which forms due to the cleavage of adduct). We have tried with different column chemistries like C8, Cyano, Diol (different make) and silica and found that Diol column (YMC /Betasil) is more suitable than other columns. The mobile phase used is Acetonitrile with 0.01% ammonium hydroxide and Isopropyl alcohol with 0.01% ammonium hydroxide in the ratio of 85:15. The detection wavelength is 254nm. The RT of analyte is about 6 min and impurity is at about 4 min. sample prepared in pure Acetonitrile.
We are facing the following problems with this method.
1. Area of the analyte peak is decreasing continuously. Some times it is working fine but many times % RSD is failing for standard injections.
2. Compound is stable in Acetonitrile, so there is no chance of the degradation. The compound is highly soluble in Acetonitrile so there are no chances for the precipitation to occur.
3. Some times the response is nil.
4. We have saturated the column by injecting the concentrated solution of the analyte by repeatedly injecting about 5-7 injections. Even then we are facing the same problem. Why the response is decreasing.5. If we inject the same sample next day, the response will come but it goes on decreasing up on injection to injection.
Please suggest some way to resolve this.