Nalizer,
The UPLC from Waters is just a LC system that can go to higher LC pressures. The advantage is that with higher pressures you can go down to stationary phase particle size and either increase your peak capacity and/or your analysis time (I can give you some theory on this if you want).
As a result you will have an advantage only if you opearate with smaller particle sizes such as the 1.7 uicro from Waters. This stationary phase has a very narrow particle size distribution so the backpressure is pretty high.
I have heard contradictory versions of one thing and this is if you will be able to operate that system with any type of other than Waters stationary phases (maybe Uwe can comment on this). Now, other companies have started manufacturing sub 2 um particles but the size is not as uniform so you can run them with conventional LC instrumentation. The Agilent 1.8 um particles is an example. Agilent claim increased efficiencies without the drawback of higher backpressures.
Part of our departement have already make the move towards higher pressure even before Waters commercialize their systems (we are currently working at 10000 psi, but we are mainly working in the area of proteomics where high peak capacity is a must.
Now, in your question if industry is moving towards this technology is too early to say but I will give my personal view in the subject and people are wellcome to agree or disagree...
I think that the rules changed about 5 years ago when electrospray mass spectrometry was accepted by the scientific community as the detector of choice for most analytical applications. Several years from now, almost all the analysis will be performed with a mass spectrometer and the rest of the detectors will be there only to supplement for specific applications. As a result what will be the future of the rest of instrumentation (including the separation science as we know it) will depend on what is more compatible with the MS and the needs of biology. The transition of analytical LC to micro and nano LC is an example of this type. The MS is much more compatible with nano flows and biologists wants to analyse as little amount of sample as possible (the future is down to single cell).
The scientific community asks for as much sensitivity as they can get, they want as large dynamic range as they can get and they want to find ways to get rid of the ion-suppression phenomena that occurs during analysis of complex matrices with MS. Dynamic range depends on somehow to your sensitivity and sensitivity is increased the lower your flow rate while due to the higher electrospray efficiency at these flow rates, ion suppression is limited if not eliminated.
Now, while ultra high pressure liquid chromatography was investigated in the first steps of HPLC it was quickly abandoned as highly challenging and believe me there are several challenges associated with it as you need to redesign everything (valves, fittings etc...). Another problem was that for conventional chromatographic columns (i.e. 4.6 um ID and up) are heating up due to water molecule friction when you exceed the 15000-20000 psi. Heating is not a problem for capillary columns.
So why higher pressures? If you are convinced that capillary columns and nano-flow systems will be the (solely) separation front of the MS then there is a bottleneck that has to be resolved:
a) Packing of capillary columns less than 50 um (and especially less than 30 um) is quite challenging even with 3 um -we were able to go down to 15 um with these particles... I assume that the packing of ultra thin capillaries will be easier with smaller and more uniform particle size. Going down to such small ID will serve the purposes of biology (very small injected amounts and high sensitivity) and the purposes of pharmaceutical industry (elimination of ion-suppression) for accurate quantitative analysis and higher throughput.
b) Chromatographic history shows us that there is a continues trend of going in lower and lower particle sizes and I do not think that this trend is going to stop but after a point it would be necessary to have higher pressure operating LC systems... It would be of course more and more difficult to make small particle size porous silica particles but maybe if you can go down to 0.1-0.5 um particles you do not need them to be porous (maybe someone can comment on this).
Finally... I think that another technology that will come in to the play and that is ion-mobility that will be an indermediate gas separation step between the LC and the mass spectrometry. Due to the peak capacity offered (or that will be offered) by ion-mobility you will need very fast, high efficient separations in the front end (LC).
Anyway, this is my point of view. people are welcome to disagree (and I hope that they will... this is a discussion forum after all). I think that 5 years from now we will have a better idea if the above will be at all true (who knows... maybe the "lab on a chip"?).