Hi all,

I have just bought a CP-Wax 57 CB, which has a maximum isothermal operating temperature of 200C and maximum programmed one of 225C. In the same GC oven I will have an HP-FFAP, on which I run an analysis with a final hold at 210C for 5 minutes. The overall analysis time on the FFAP is 25 minutes and I can run up to 15 samples a day (although not every day).

As far as I understand, the Wax can be taken for a short time at 225C but I am not sure of how frequently this can be done before damage to the phase occurs. Can the Wax sustain short but repeated exposures at 210C or do I really need to stick to 200C?

Looking forward to hearing your opinions.
My opinion/guess is you are OK taking the Wax column repeatedly to 210 C for 5 min since the max temp is 225 C. I would make sure that in my FFAP method that I have carrier flow going through my Wax column.
That was my understanding too but as it is a new column, I wanted to double check before I do something daft. I will make sure that there is enough flow going through.

Many thanks!
3 posts Page 1 of 1

Who is online

In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry