padtwo » Tue Dec 16, 2008 8:17 pm
Thank you for the replies so far!
Buffers are 0.1% Formic in water for buffer A, and 0.1% Formic in ACN for buffer B
Column prep would be washing the column with 100% B for 5-10 minutes, then equilibrating it with 90% A and 10% B for 10 minutes or until the baseline looks ok.
Program then runs for 5 minutes at 90%A after loading then the conc of B increases from 10% to 100% over 15 minutes, then continues at 100% for a further 2 minutes to make sure everything is off. A total run of 22 minutes.
I know the 214 is close to the limits, elsewhere we use 280, but we've been ok in the past and this is a relatively new problem. Machine has a DAD which I thought was one of the better detectors, hence getting away with reading near the limits, but I'm quite new to this and would welcome any advice!