You should absolutely set up and use a degasser with your HPLC system. Degassing ALL mobile phases types is beneficial to reducing pump cavitation and check valves from getting stuck (from air bubbles). Helium sparging (degassing) is superior to all degassing techniques because it both displaces most of the gas from the solution in a continuous manner and is chemically compatible with all mobile phase types (which inline vacuum degassers are not). Their are only two negatives associated with using high purity helium to degass the liquids. Helium sparging is inconvenient due to the heavy gas cylinders which must be exchanged every few months and the gas is expense.
Setting up continuous helium sparging is very simple (never set it up for ON and OFF use as that defeats the purpose!). Use a secondary pressure regulator off of the high pressure tank and set it for about 3-4 psi. Run the gas to SS or Teflon perforated solvent pickup frits for delivery to the MP bottles and make sure you position them higher than and away from the MP solvent pickups (to prevent them from sucking gas into the system). You will end up with a mobile phase that is thoroughly degassed and has the lowest oxygen levels (great for low UV work).
Inline vacuum degassers were created for convenience, but there are downsides to their use too. They do not provide the level of degassing that helium sparging does (few need that level), but if setup and used correctly will work great for many applications. Limited chemical compatibility of the vacuum chambers and the fact that they must be regularly flushed (wastes solvent) are two drawbacks which must be attended to. Also, never leave a degassing channel empty, always run with them ALL filled with liquid (per the manufacturer's spec). They also wear out and degrade inside over time so need to be serviced every 4 to 5 years (or your mobile phase may become contaminated).
Based on your request: If you want convenience and your mobile phase solutions are compatible with the vacuum degasser you selected, then use it. The easy plug-in convenience is great and if properly serviced will provide good degassing improving your baselines and freeing you from spending all of your time getting the pump to work. If you have easy access to high purity helium, then sparging with He will work with everything well and you will have no need to buy or use a vacuum degasser.
Here is a link where it was discussed:
https://www.researchgate.net/post/What_ ... g_for_HPLC