I am in the process of replacing a 0.25mm column with a 0.32mm one in my FID. Since I installed the new one yesterday, the detector is having some ignition issues. When I first tried to restart it, it failed (the H2 flow was reading 50ml/min at that point, above the 40 set value). With the H2 and O2 flows off (kept the make up on by mistake), I noticed that the H2 flow was still high (above 30ml/min), which made me think that the jet was blocked. With all flows off on the FID but carrier at 0.5ml/min, the H2 actual flow displays around 0.3-0.4ml/min.The FID had been running without any issues until I swapped the columns yesterday.
Wondered whether I had installed the column too far or whether the wider ID was the issue but I have tried today to switch back to a 0.25mm and still seeing the issue. Now the FID made the popping sound a few times in a row, came on and switched off again. It looks like a flow issue but don't know whether it is due to jet or my poor installation. I am waiting for some cleaning supplies, as well as a new jet but wondered whether anybody would have any ideas of what could cause this?
GC7890A
Flows: H2: 40ml/min; O2: 400ml/min; make-up: 25ml/min
Optimised capillary fitting-installed column by inserting it until it bottoms out, then retract by ~1mm.