LC-MS/MS system for organometallic compounds
Posted: Fri Nov 21, 2008 5:40 am
by tswapna
At present I am working in a reserach lab wherein we are sysnthesizing organometallic compounds ( mainly with transition metals ).
We want to use ESI based LC-MS/MS system to do quantitation analysis
of our compounds.
Can we use LC-MS/MS system for our application?.
Can any body explain me in detail.
Thanks in advance
Posted: Fri Nov 21, 2008 11:52 am
by bhuvfe
tswapna,
It will depend on the stability of your organometallics.
If they are labile you may need cold-ESI see link:
http://www.jeolusa.com/DesktopModules/B ... attachment
Otherwise a standard ESI will do ( see link:
http://dissertations.ub.rug.nl/FILES/fa ... /04_c4.pdf)
Regards,
bhuvfe
Posted: Fri Nov 21, 2008 4:19 pm
by lmh
I've had some that have worked beautifully, and others that have been disastrous, so all I'd suggest, is try before you buy an instrument!
To be fair, I've only tried qualititative work with them; as you know, ESI is not an especially quantitative method without good standards.
If you're lucky, depending on transition metal, you can get some beautiful isotope peaks, and if you want to establish how many metal atoms are present, it can be a good technique.
As bhuvfe said, temperature matters. I've used an elderly Thermo ion-trap system where the default capillary temperature is 350 degrees, but for some organometallics, it's been necessary to reduce this as far as 200 degrees. Probably many others wouldn't even survive this.
Posted: Mon Nov 24, 2008 3:55 pm
by jerole
Do you have an ICP-MS? It is very easy to couple LC to an ICP-MS for quantitative analysis of your organometallic compounds.