sensitivity of Agilent MSD vs. Varian Saturn in MS/MS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
			3 posts
							Page 1 of 1
					
	
						I am working on an application looking for ultra-trace (< 1ppb) concentrations in fish tissue of a series of PAHs and their partially saturated analogs.  It appears as though I need at least an order of magnitude more sensitivity than I currently achieve with my Agilent 5973 MSD in SIM mode.  I also have access to a Varian Saturn 2000 Ion Trap and was wondering if I can get greater sensitivity using MS/MS.
					
									- 
																									  
- CE Instruments
						Simple answer is no, more complex answer is possibly.  A trap is usually more sensitive in full scan than a quad, the trap can be more sensitive on a single ion , Quads are usually better on SIM (more than 1 ion).  The best way to improve sensitivity is to get more on column   Can you improve you sample pre treatment, a better injection method/ protocol ?
   Can you improve you sample pre treatment, a better injection method/ protocol ?
If you only look at MS detection, SIM multi ion is best on a quad. Except where you can trap a single strong ion using a trap and carry out MS/MS to confirm identification. It is also possible with quad MS/MS to use the primary quad to remove interferences and to enhance signal to noise for detection and identification.
									 Can you improve you sample pre treatment, a better injection method/ protocol ?
   Can you improve you sample pre treatment, a better injection method/ protocol ?If you only look at MS detection, SIM multi ion is best on a quad. Except where you can trap a single strong ion using a trap and carry out MS/MS to confirm identification. It is also possible with quad MS/MS to use the primary quad to remove interferences and to enhance signal to noise for detection and identification.
- 
								
 
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
						Since you are looking at PAH's, CE's comment about more on column is a great one.  Look at large volume injection to greatly increase the injection volume.  
Best regards.
									Best regards.
			3 posts
							Page 1 of 1
					
	Who is online
				In total there are 36 users online :: 1 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: Google Adsense [Bot] and 35 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Adsense [Bot] and 35 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science