Troobleshooting LCMS THC-COOH in urine
10 posts Page 1 of 1
Hi
My first post.
Hope someone can give me some tips on issue i have with LCMS analysis of THC-COOH in urine.
I am doing alkaline hydrolysis with ammoniumhydroxide 25% (60uL amhyd, 120 uL IS (THCA-d9 in water), 240 uL urine or spiked urine calibrator). So no SPE, dilute and shoot method.
I have done ion suppression and I had 40% suppression compaird to blank prepparation (ie water instead of urine) and THCA and THCAd9 behaves similar.
However when i spike THCA and IS to urine and check revovery agains blank prepparation THCA and IS do not behave similar between urines from different co-workes (subj healthy, meds and no meds, not using cannabis).
For exampel matrix X IS respons down 41% THCA only 17% and matrix P IS down 34% and THCA 27%.
Compairing the different urine matrixes to each other shows that for one matrix the IS can be higher but THCA can be lower than for an other urine matrix.
This affects the slope of my calibration curve. I have noted a slope difference of 30% which off course will affect my results proportionally and I get very high LOT to LOT variations.
Has anyone seen this effect or have any tips or ideas that could help me? SPE off cours could be one tip but trying to sett up method performed at several labs in Sweden.
Appologies for any misspellings or grammar errors since English is not my first language.
Sincerly
Pär
Welcome to the forum.
I don't have a lot of experience with urine analysis, but I can imagine differences in urine composition from different persons. I am not sure if you are looking at suppression or recovery differences here. If you want to look at suppression only, you should look at the differences in signal when the analytes / IS are spiked AFTER the sample preparation for both a blank and a matrix sample.
For a matrix effect, matrix has to co-elute. That's why an IS co-eluting with the analyte is so important in LCMS to correct for matrix effects. Your IS has 9 deuteriums, and the more deuteriums, the more tendency to elute earlier in RPLC than the corresponding analyte. If this retention difference is significant, this can explain the 30% difference in suppression between the analyte and the IS. I saw this behavior before using an IS with 8 deuteriums that was only 50% overlapping with the analyte. You could try to tweak the LC run to make them co-elute. Alternatively, if you have the sensitivity, you dilute everything (or inject a lower volume).
This seems to be a good reference for what you're doing.
I don't know for sure how deuterium behaves in alkaline solutions but I have had times in acidic solutions where the deuterium will be replaced by hydrogen so the recovery of the native analyte looks higher and the deuterated analog is low.
What purpose does the sodium hydroxide serve in the preparation?
Hi Rndirk
and thanks a lot for your tips.
With ion suppression I actually meant post column infusion. Sorry for the poor details.
You are completely right. Of course I should have spiked post preparation so no effects from the prepp it self would confuse me. I´ll do that straight away. Stupid of me.
Actually your previous experience is valid also for me. I don’t have 100% coelution of THCA and THCAd9. Tried to calculate it but tricky. Have approx. 70% overlapping area.
I`ll try your tips and read the article.
Hi James
and thanks for replying.
I´ll look in to this. I have also had some concerns of what’s happening in the vial.
There are several articles on alkaline hydrolysis of THCA-glucoronide.
I ‘am not using sodium hydroxide but ammonium hydroxide.
It has to be added to hydrolyze the THCA-glucuronide excreted in urine back to THCA. It is the first analyt of choice since it was historically hard to access THCA- glucuronide standards I think. Cut of values are based on THCA.
Hi James
Makes sense to me now, thanks.
Would it be possible to capture the target analytes on a solid phase extraction cartridge from the extract, wash with water to remove most of the matrix then elute with solvent for a cleaner sample to inject?
Hi James
Yes absolutely SPE would be possible and there are applications on this. But at this time we are trying to get dilute and shoot working since others are doing it and we want to keep it simple. But of course if it just won’t work we will shift focus to SPE.
If alkaline hydrolysis would be the problem (deuterium - hydrogen exchange like James described), there's an alternative (and milder) method to hydrolyze glucuronide metabolites using enzymes. Here's some information.
I would suggest to start by looking if it's the sample preparation that's giving recovery differences, or it's really matrix effects due to co-elution with matrix.
One more thing: the article I mentioned earlier neutralizes the alkaline extract and adds organic solvent to precipitate and get rid of a bunch of stuff. Is this something that you do as well, or do you just inject the alkaline solution?
Hi Rndirk
Unfortunately I have not been able to download the article yet. But in my case it just dilute and shoot and no neutralization or precipitation.
The strange thing is that the method is working for others and has been published by Beck et al (J of Chrom B 871 (2008) 101-108).
I am looking in to the preparation part as we speak.
Among other things I have observed that when adding 60uL 25% amhydroxide + 120 uL THCAd9 in water + 240uL spiked urine (10% MeOH) i get a cloudy solution almost straight away. But when substituting THCAd9 with 120 uL water I get less precipitation and not straight away.
I’ll continue getting access to the articles you mentioned.
Hi all
just wanted to post the solution to my problems with THC-COOH and d9THC-COOH not behaving similar in matrix. Perhaps it could help someone else.
After recommendations from Rndirk i performed different kind of addition trials both pre och post with analyte and IS.
I then noticed that the THCA/d9THCA ratio was constant when spiking directly to my vial.
The difference between what i did at first and the spiking exercises was that when doing "normal" curve i prediluted my THCA-methanolsolution in negative urine before adding the spiked urine portion to the vial. When i started to add THCA-methanolsolution directly to vial the issue was solved.
So it turned out that the adsoprtion of THCA to the walls of the tube i used for dilution differed between different matrixes.
Thanks for all help.
Best regards
Pär
Users browsing this forum: Semrush [Bot] and 22 guests