Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I am analyzing iohexol with isocatic (MeOH/H2O/acetic acid) using 1100 Agilent (DAD) at 254nm, phenomenex Luna column. For high concentraion of standard, good peak at 11.6min. But for low concentration standard, got negative peak at 12.2min. I used all LC grade of solvent. acetic acid is ACS grade (but used another bottle with LC/MS grade of acetic acid, same results). Ran blank, pretty clean and no peak. Look for advice why. Thanks.

One possibility is that something is in your sample (or blank, for that matter) that absorbs more strongly at the reference wavelength than it does at 254.

Check to see if the reference wavelength is set on the DAD. If it is, turn it off and see what happens.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

thanks. Tom. Tried that and the negative peak is still there. Developed another method using new column and mobile phases, now the negative peak is gone.

I'm glad you got it fixed. I'll go out on a limb and speculate that you had a "system peak" with just enough of an RI shift to show up (but thats a very long limb! :wink: ).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 3 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot], John Guajardo and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry