by
krickos » Thu Nov 06, 2008 11:00 am
Hi
Given the information I am putting my money on contamination of the valve/loop
One issue with Agilents valve/loop solution is that is it sensitive for contamination especially for polar solvents that can cause poor reproducibility and loss of sensitivity.
Valve/loop temp should be equal to or exceed the highest bp of the solvents including sample sample solvent (Water, DMF DMSO etc)
If you have been runiing samples on the instrument dissolved in DMAA, DMF, DMSO, or DMI with this method or previous other method with these valve/loop temperature setting, you WILL build up contamination in the valve/loop over time.
I have seen this so many times, had the valve/loop temp been like 50-55°C you would also seen it on methanol but now the valve temp exceeds the bp of methanol with a margin. I have seen the exact problem with ethanol on similar parameters on several instruments on our sites. And it almost always boils down to a contamination issue.
The down side with temperature setting in Ph Eur and USP is that they are most likely derived from methods with a heaspace using flow injection rather than a loop thus less sensative for contamination.
If running the Ph Eur or for that matter USP general methods on an Agilent sytem with loop/valve please follow my advice about temperature in the quote, and you will see less contamination issues and need less frequent steam cleaning or manual cleaning of loops.