I have gotten word that our raw material qualification group simply can not get this assay to work. So, I took a look at it. It seems no matter what I do, I either consistently fail peak area ratio %RSD, peak tailing, or a combination of both. The peak area of the ISTD is consistent from injection to injection, the peak area of the cholesterol is not.
The method is being run exactly as stated in the EP and we are using a Restex Rtx-1 column (though I have also tried with a DB-1 and what ever Supelco's version of the XX-1 columns are. All produced similar results). I have tried the assay using a regular split injection linter, a focusing injection liner, siltek coated split injection liners; liners packed with silanized glass wool as well as quartz wool; modification of injection port temperature by 10% to increase reproducibility.
As you can probably guess I am operating under the assumption that, due to the fact that only the cholesterol peak area is changing, the cholesterol must be in some way reacting with the liner.
Any ideas as to how to fix the problem would be greatly appreciated!