Hello,
I need clarification regarding the variation of oganic composition of mobile phase for checking the robustness of a method. In the case of isocratic method the organic composition is varied by 10% (plus or minus) from the orginal value. My question is how the same effect can by studied in the case of gradient elution say for eg: M.phase A: 0.1% TFA and M.phase B: Acetonitrile. "Whether the study of gradient robustness is enough and whether it gives the effect as varying the Organic composition".
Thank you.
Santosh Gandhi.

First of all, plus and minus 10% is a HUGE change. More typical values for a robustness study would be plus and 1 or 2% (absolute) from the specified value. Assuming small molecules with an S-value of 2.5, the gradient equivalent would be about a 10% change in gradient time.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 20 users online :: 3 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry