XL » Sun Oct 19, 2008 3:10 am
													
 
					
						Benzalkonium quats can be lost by absorption during sample preparation (e.g. filter materials) and autosampler vials by hydrophobic interaction (plastic vial) and ion-exchange (glass vial). I prefer to use glass vials and pre-treat the sample vials with a more hydrophobic quat that is not the analyte of interest to mask the silanol groups. Also inject such masking solution to the LC system to mask adsorption sites in the flow path before analysis. As a result, the recovery can be improved.
For analysis of benzalkonium salts, UV detection should provide quite good detection limit (low ppms) although sometimes ELSD is sufficient for higher concentration samples (>10 ppm). However, very often the analysis can be interfered by other components present in the sample. Therefore, to achieve better sensitive (ppb) and selective detection, suppressed conductivity detection is the solution in which case all neutrals and anionics are transparent.
Dionex is offering a specialty column, Acclaim Surfactant, designed for analysis of cationic, anionic, amphoteric and nonionic surfactant analysis. The following link contains some information on cationic surfactants analysis including benzalkonium salts. Please pay attention to Figures 1, 2, and 6-9.
http://www.dionex.com/en-us/webdocs/685 ... 01-ACS.pdf
Please contact me if you need further details.