Data review
Posted: Sun Oct 12, 2008 4:50 pm
by mini
I have a question, My run went through every injection of table set. I checked every vial it has have been injected, but when I review it only have haft of data. Does any know what's problem? Thanks for your time
Mini
Posted: Sun Oct 12, 2008 5:04 pm
by tom jupille
We don't even know what kind of instrument and control software. :roll:
Before any one can provide advice, you have to provide details about your problem.
Posted: Sun Oct 12, 2008 5:11 pm
by mini
I use HPLC Water 2695 and Empower software
Re: Data review
Posted: Tue Oct 14, 2008 12:52 pm
by JGK
Mini
I assume:
1. Your printer didn't run out of paper.
2. You didn't specify 1 injection/vial instead of 2.
3. Only half of your sequence was injected using "run and report"
Were you doing Duplicate injections from the vials? if yes are you reporting vial summaries only.
Have you checked the results for this sequence? Are all the results present? if yes reprint the results, if not you need to look closely at your injection sequence.
Posted: Tue Oct 14, 2008 5:28 pm
by DR
Data may be buffered in the LAC/E - have your Empower sys. admin look into that (or at least get permission before rebooting the LAC/E to force a data transfer).
Your sys admin will appreciate it if you provide the exact name of the project and system used to gather the data that appears to be missing.
If your data was collected via SAT/IN and some is missing, check the message center under the "All" or "Acquisition" tabs to see if there are any red lines time stamped around when you were collecting data...
Posted: Thu Nov 13, 2008 7:38 pm
by Jade.Barker
I have Waters/ Empower 2. I am pretty new, so this might be too elementary:
1. Did you hit the Update button?
2. Is it possible you cleared the calibration curve and it isn't calculating?
Don't laugh - it can happen :oops:
Posted: Wed Apr 01, 2009 8:29 am
by larryaducady
Couldn't it be possible that your Project ran out of tablespace?
Posted: Fri Jul 31, 2009 11:04 pm
by cdsman1001
Very much possible