What is the difference between single quad vs. triple quad mass analyzer?
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
Single quad is single quad. Triple quad is 3 quad set Q1, Q2 and Q3, in which Q2 is in fact a CID, only Q1 & Q3 are mass filters.
-
- Posts: 199
- Joined: Tue Feb 08, 2005 5:27 pm
Hello; What thohry says is correct; but I want to add somthing else.
Usually what we call Triple quads or Tandem systems is not alway three quad puted togeter; so instead to have a quad as collition cell some companies are using other "diveces as T-wave or Hexapoles for this.
The diference betwen single quads antr "triple" quads is:
Single Quad; you ionize your molecule and then you can filter or select one m/Z ration using your quadrupole; and then detect it. or get all the fragmentation spectra with out filtering elting pass all the aion to the detector
Triple Quad:
1) you ionize and use only your first quad.
2) You ainoze, and filter one m/z ion in yor first quad and fragment it again in the "second" quad and obtein a new fragmentation spectra from an ion.
3)You ainoze, and filter one m/z ion in yor first quad and fragment it again in the "second" quad Then you select a new m/Z in your thrid Quad.
Usually what we call Triple quads or Tandem systems is not alway three quad puted togeter; so instead to have a quad as collition cell some companies are using other "diveces as T-wave or Hexapoles for this.
The diference betwen single quads antr "triple" quads is:
Single Quad; you ionize your molecule and then you can filter or select one m/Z ration using your quadrupole; and then detect it. or get all the fragmentation spectra with out filtering elting pass all the aion to the detector
Triple Quad:
1) you ionize and use only your first quad.
2) You ainoze, and filter one m/z ion in yor first quad and fragment it again in the "second" quad and obtein a new fragmentation spectra from an ion.
3)You ainoze, and filter one m/z ion in yor first quad and fragment it again in the "second" quad Then you select a new m/Z in your thrid Quad.
-
- Posts: 1894
- Joined: Fri Aug 08, 2008 11:54 am
try doing a Google search for mass spec and "precursor ion scanning" and "neutral loss"; these are two very interesting ways to use a triple quad that have been around for a very long time. Also try looking up "MRM" methods, which need a triple quad.
-
- Posts: 247
- Joined: Sat Apr 28, 2007 3:24 pm
Also, triple quads are typically much more sensitive than single quads due to their capability of doing MRM scans, etc.
-
- Posts: 199
- Joined: Tue Feb 08, 2005 5:27 pm
I have seen nice videos in youtube
just typing "mass spectrometry"
6 posts
Page 1 of 1
Who is online
In total there are 223 users online :: 0 registered, 0 hidden and 223 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 223 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 223 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science