We have both LC/MS and GC/MS in our lab. Power failure is not common, but it happens. I think that it should do harm to the systems, but what can be affected? Please help.
thanks
We have both LC/MS and GC/MS in our lab. Power failure is not common, but it happens. I think that it should do harm to the systems, but what can be affected? Please help.
thanks
2 uncommon things that I know of:

1. When the rough pump goes down if the MSD source/quads stay really hot you have the potential to suck oil vapors from the rough pump into the source chamber.

2. We've had a circuit board die when an instrument was left on during a power failure and a surge went through the lines when power was restored.

Does you lab have an emergency power supply (diesel generator etc)?

between power failure and emergency generator activalion any data aquisition will stop.

You also run the risk of power surges when the generators "kick in" or when you switch back to mains power.

For this type of system, our lab has each machine equiped with a UPS unit (with current smoothing) to ensure data acquisition is unaffected by mains interruption.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.

It all depends on which system you have or will buy, some vendors have anti-suck up valves in them so that fumes don't back up into the Vacuum chamber...some don't....you will have to check.

It's always a good Idea to have these expensive systems on a UPS (un-interuupted power supply) ...so you don't blow boards...it also avoids power spikes and the like.

The UPS is the best for maintaining up-time. But the cost is up there too. An alternative is to be sure that your rough pump has an anti-back streaming valve and to use a power conditioner. The power conditioner is not cheap either, but costs less than the UPS.

Even without the lights going out, spikes on the line eat instruments. Some locations have "dirty" power and the addition of a line conditioner reduces the number of service calls made.

Some kind of protection on the pump and line are dictated by the value of your equipment. The additional step of going to a UPS is dictated by the time value of your data.
5 posts Page 1 of 1

Who is online

In total there are 273 users online :: 0 registered, 0 hidden and 273 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 273 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry