Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I have a standard that appears to be losing peak area over time. Multiple injections from the same vial show a downward trend of about 2% over a 24 hour period. A freshly loaded vial gives a response comparable to the original injection of the standard. The solution was stored at 5°C and the autosampler was also set at 5°C, so both were stored at the same condition other than the container (volumetric flask vs sample vial).

This may also be happening with the sample solutions, but I don't have the repetitive injections of a sample.

Does anyone have any experience with interactions between vials and sample solutions? Any input is appreciated.

Thanks
Ben

Interaction between your vial and your sample is a possibility. You may want to change to plastic ones at the place of glass or the other way round.

Other possibilities can be, degradation of your compound.

The more your sample is diluted the more you may notice such problems.
2 posts Page 1 of 1

Who is online

In total there are 25 users online :: 2 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry