we are dealing with a protein mixture (MW: between 30 and 7 kDa) isolated from serum, whose major component is a protein that contains 243 AA. We have used a gradient method with a RP-18 endcapped column (Chomolith, 5µm, 4,6*100 mm) to separate the protein mixture.
The result is quite miserable: particularly the peak shape of the major protein (eluted at 21-24 min, at 22 min maximal, distorted, brodening, tailing Peakshap). We thought that the major peak should be somehow heterogeneous, perhaps due to the isoforms of this major protein. Furthermore, the baseline drift is significantly variable and we can't achieve the baseline separation for many peaks.
We have tried a serial connection of two columns, decrease the flow rate (from 1.2 ml/min to 0.5 ml/min), or increase the temperature (from 37°C to 45°C). But unfortunally they diden't work.
Can anyone give us some advices, how to improve this seperation? If the column is not suit for it, which columns sould be helpful?
A small summary of the separation method:
Mobile Phase: A: 0.1% Trifluoroacetic Acid in water / B: 0.1% Trifluoroacetic Acid in Acetonitrile
Gradient: (Time,%B),(0, 28 ),(30, 58 )
Detector: UV at 214 nm
Flow rate: 1.2mL/min
Column Temp: 37°C
Column: Chromolith C18e 300 Å, 5 µm, 4.6*100 mm
Sample solvent: 3 M Guanidinium-HCl
HPLC equipment: Hitachi with autosampler (7°C)
Thanks a lot in advance