Calbration of OV-1 column for C2-C6 hydrocarbons

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello Experts!

I am a fuel cell guy working on fuel reformation, principally reforming HCs and alcohols such as methanol, ultra-low sulfur diesel, and propane. I am frequently interested in knowing the intermediate and final product composition of said reformate.

I have an Agilent MicroGC 3000 with four columns and one inlet.

Column 1 is a molsieve 5A, 12 um film, 320 um ID, 10 m length, TCD, Carrier gas is Argon. Has a PLOT U pre-column 30um/320um/3m. I use this one only to catch Hydrogen. My current method under development runs this column so:
  • injector temperature 100C
    column temperature 100C
    inject time 10ms
    run time 240s
    column pressure 30psi
    backflush 10s
Column 2 is the same, with Helium as the carrier, same pre-column, method as such:
  • injector temperature 100C
    column temperature 90C
    inject time 100ms
    run time 240s
    column pressure 30psi
    backflush 10s
This is my column for oxygen, nitrogen, methane, CO.

Column 3 is a PLOTQ 10um/320um/8m, TCD, He is the carrier. His method is
  • injector temperature 100C
    column temperature 50C
    inject time 100ms
    run time 240s
    column pressure 30psi
    backflush 10s
I calibrate him for CO2, ethylene and ethane.

Column 4 is my problem child. He is a new OV-1, 1.2um/150um/8m, TCD, He is the carrier. His method is:
  • injector temperature 100C
    column temperature 70C
    inject time 10ms
    run time 240s
    column pressure 30psi
    backflush 10s
I intend for this one to catch (I have known certified samples of):
  • 1,2-propadiene
    1,3-butadiene
    and hexane
in one set of samples,

and
  • propane
    butane
    benzene
    and toluene
in the other set of samples.

I am having trouble establishing what peak is what.
Could someone give me a possible clue of what the expected elution order and times might be?

Infinitely in your debt in advance,

wave_man

I run on HP5-MS column, the orders of the 2 set of samples are exactly what you ordered. For OV-1 which is similar to HP1, the order should be the same.

I run on HP5-MS column, the orders of the 2 set of samples are exactly what you ordered. For OV-1 which is similar to HP1, the order should be the same.
Thank you kindly for your reply. I will try to post some chromatograms to show my difficulties. I think I have two many gases mixed together for an "easy" run through the calibration process. I have started from the simplest ones, with only stationary gases (is that the correct term?) so that I can see if they are influencing the OV-1 output at all. I believe nitrogen has a big change on the shape of the composite (garbage) peak of the OV-1. If folks are interested, I will post my 'grams. If it is considered rude to post pictures, i will not...

wave_man
3 posts Page 1 of 1

Who is online

In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry