Ballast on rough pump in LC/MS
Posted: Thu Sep 11, 2008 12:36 pm
by haiedc
Hi all,
Why đo I have to open the ballast on rough pump in LC/MS, while not in GC/MS?
Thanks.
Posted: Fri Sep 12, 2008 10:27 am
by lmh
In a GC, the vacuum pumps are removing a fairly small trickle of gas. In an LC-MS, they are removing a larger volume of vapour from a much less volatile liquid (the hplc solvent). The vacuum pumps therefore tend to accumulate solvent in their oil in the LC system, which isn't a problem in the GC system. The ballasting allows them to get rid of accumulated solvent, and keeps the oil cleaner.
Posted: Wed Oct 15, 2008 9:15 am
by sindhu
Hi,
Whatever Imh has told you is true