Ballast on rough pump in LC/MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi all,

Why đo I have to open the ballast on rough pump in LC/MS, while not in GC/MS?

Thanks.

In a GC, the vacuum pumps are removing a fairly small trickle of gas. In an LC-MS, they are removing a larger volume of vapour from a much less volatile liquid (the hplc solvent). The vacuum pumps therefore tend to accumulate solvent in their oil in the LC system, which isn't a problem in the GC system. The ballasting allows them to get rid of accumulated solvent, and keeps the oil cleaner.

Hi,

Whatever Imh has told you is true