Silver is one of the worst elements to keep stable for certain. The EPA methods will suggest that anything over 100ppb can fall out of solution, personally I see it begin to fall out at 250ppb. The first day the calibration is good up to 1ppm, if I run those standards on consecutive days the 1ppm, 500ppb and 250ppb will begin to show lower recoveries, the 100ppb will usually be fairly stable. We just stop our calibration at 100ppb now. Higher HCl can help but that interferes with other elements so unless you are only looking for Silver it isn't worth it.Silver Chloride is slightly soluble in nitric acid. Each time you dilute you get a slight amount more (taking DF into account).
For 20+ years I've run ICP-AES and ICP-MS using a 2%HNO3-1%HCl final dilution matrix. I do my digestions with those acids in that ratio. This will keep at least up to 1 ppm Ag soluble at the instrument in standards and samples. That way chloride interference is not a problem. You also get better recoveries of As, Sb, and Tl in complex matrix with chloride present.