I've had success using the fused-core column for impurity analysis of small molecule pharmaceutical compound. The drug substance has 3 aromatic rings, and 2 cholo atoms on different rings. One Cl is switched from position 2 to 5, resulting in a regio-isomer that’s very similar to the parent compound.
We screened more than a dozen different types of columns, with acetonitrile or methanol as the mobile phase B. This Cl-isomer could only be separated with methanol and C18 column. The impurity peak eluted at the tail of the main peak. Temperature and pH didn’t do much.
We compared Zorbax SB-C18 (1.8 micron) with Halo C18 (2.7 micron), both 50x4.6 mm, flow rate at 1 mL/min. back pressure on the Zorbax was 150 bar with resolution 2.7, while the Halo only gave 84 bar with resolution 3.2. the resolution here is a for a different critical pair, the Cl-isomer couldn’t be baseline separated with such short columns. We also compared Luna C18 (3 micron), which resulted in slightly lower back pressure but worse resolution than Zorbax. In both comparisons, peaks came out slightly earlier on the Halo column, the elution order remained the same. Halo just had the extra efficiency we needed without getting over the pressure limit of our instruments (400 bar).
In the end, we used the 150 mm long Halo C18 column to resolve the Cl-isomer (resolution 2.3 to 2.7 on different instruments/columns). In order to achieve desired sensitivity for impurities (LLOQ 0.05% of the main peak), the injection volume was 25 uL at 0.3 mg/mL. This resulted in a slightly fronting peak (UPS tailing factor 0.9). This may be an indication of the reduced sample capacity of the fused core particles. This method has been validated, used for GMP lot release and stability testing, and transferred to a contract manufacturer.
Overall, I am happy with our experience with the Halo C18 column (also sold as Ascentis Express C18).