RP-HPLC retention problem??

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi

I have a peculiar problem in my RP-HPLC, I am using a RP-HPLC method for quantification purpose. I am using C18,4.6 X 50mm, 5µ, 300 Å (discovery column)…every time I run a fresh sequence in my first 3-4 injection the sample doesn’t show any retention (my sample elutes with void volume peak..) after 5th or 6th injection the sample recovers its retention (around 5.0min)..this happens every time I run a fresh sequence..(I am doing proper storing of my column ..60% ACN)…

Could some one tell me the reason for such a phenomena?

Thanks a lot

You should give some more details about the method but my first guess would be inapropriate equilibration of the column (i.e. you need to equilibrate your column longer before you start your sequence). We can be more precise if you give some more info about the method (i.e. mobile phase, equilibration time, mobile phase, isocratic or gradient etc...).
2 posts Page 1 of 1

Who is online

In total there are 22 users online :: 3 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry