HP 1090 and Agilent 1100 FLD
Posted: Mon Aug 11, 2008 5:25 pm
by guttridm
I'm currently trying to hook up an Agilent 1100 series FLD to the HP 1090 HPLC machine we have in the lab. The FLD is installed and recognised on the chemstation and all of its parameters can be set during the Edit Entire Method menu. The FLD also shows no Error messages and is ready to run. However, it wont actually run a scan during the HPLC sequence! How can this be done? The only way we are currently able to run a fluorescence scan is when the HPLC is not running a method or sequence. Is it possible to run an FLD scan in sync with the HPLC? If so, how is this done?
Posted: Mon Aug 11, 2008 5:55 pm
by Consumer Products Guy
Did you also connect a remote cable, in addition to the signal cable? Are you sure it's not just software, like the FLD sognal not in "Signal Details"?
Posted: Wed Aug 13, 2008 9:07 am
by guttridm
I was maybe thinking it was some sort of software issue on the Chemstation, such as box I have to tick or something along those lines but having searched each menu and subsequent dialog boxes, I can't find anything!
Posted: Sat Aug 16, 2008 8:26 pm
by giacomo56
You will need to put a stop on the HPLC run, in the sequence table method, so that the peak of interest will be in the FLD cell so that the monochrometer of the FLD can do the scan, then complete the run....pretty hard to do a scan while a peak is moving thru the flow cell.