Calibration of an Agilent 6850 GC using Standard Gases

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
Hi everyone,

I am using a Agilent 6850 Network GC with a splitless injector to analysis for residue ethylene oxide testing. I currently calibrate my GC using a range of ethylene oxide in water standards (1ppm - 100ppm). These standards are made according to ISO 10993-7. I have purchased a range of ethylene oxide gas standards in Nitrogen at between 1ppm - 100ppm. I am having problems with the calibration of my GC using the new gas standards. I can't get good linearity, my results are all over the shop is it something to do with the injection volume or should the injector be a split and if so what should the ratio be.

All and any help is welcome.

Are you manually injecting using a gas-tight syringe?

In my experience, manual injection of gas stds is highly dependent on technique and developing a uniform injection action greatly improves results. Also NEVER vary the injection volume , always inject a uniform volume.

I've not had a lot of experience with gas injection valves, so I can't help if that's how your injecting your samples.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.

Hi JGK,

I am manually injecting the standards. Should I use a split or splitless injection??

Best headspace results are made using an automated headspace system. That said: you may need to make your injections more consistent, or live with your results.

Phardy,

It has been my experience that doing gas syringe injections, especially larger volumes, with a splitless injector (or packed port) causes all sorts of problems. When the 6850 goes splitless it essentially acts like a packed port.

Thus, I would suggest you either cut down to a very small injection volume or introduce a small split flow (2:1 or 4:1 perhaps.) With a small split flow, the pneumatics configuration should be able to handle the pressure pulse of a syringe injection a bit better.

Also, keep in mind that your gas standard is probably mole/mole (most are) so you probably won't get an exact match between water concentration (which is usually mass/volume) and gas standard.

Best regards.
Hi, I have some the same problem, can you said me how did you to resolve the problem
Which parameter did you using to improve the anslysis

Thankyou

Gustavo
6 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry