detector range 5973 vs 5975
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
			4 posts
							Page 1 of 1
					
	
						Just a quick question. On my 5973 I am trying to fix I notice the detector maxes out about 5-10E6 counts. On my 5975 the detector maxes out at 3.5E7 counts. Is that normal?
					
									- 
								
 - Posts: 693
 - Joined: Mon Mar 28, 2005 5:38 pm
 
						They each came in several flavors and with different software so I would expect  different clipping thresholds.
					
									- 
								
 - Posts: 3477
 - Joined: Mon Jan 07, 2013 8:54 pm
 
						I have had E^7 counts before on 5973's so could be software/firmware versions or possibly EM volt settings, the higher you have to run the EM volts the less linear dynamic range you will have.
I like to try an keep my normal calibration range at a max peak height in the 1 million range just so when I have high junk in the background I still have good peaks for the targets without far of clipping right when a target elutes.
									I like to try an keep my normal calibration range at a max peak height in the 1 million range just so when I have high junk in the background I still have good peaks for the targets without far of clipping right when a target elutes.
The past is there to guide us into the future, not to dwell in.
				- 
								
 - Posts: 1113
 - Joined: Wed Dec 22, 2010 10:17 pm
 
						Yea I'm running it a lot this week and the sensitivity seems pretty similar to my 5975 which has a very old (OEM 12 years) detector that runs at 1900ev. The difference in scale kind of threw me off while I was diagnosing it but reviewing old data that is normal for that instrument.
					
									
			4 posts
							Page 1 of 1
					
	Who is online
				In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: Baidu [Spider], Semrush [Bot] and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Semrush [Bot] and 17 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 