What do you guys do for the impurity syst suit with regards the sensitivity? Do you do:

a) LOD only

b) LOQ only

c) both LOD & LOQ

d) neither :o

We use the LOQ for imp's, but LOD for the cleandown analysis. What is the justification for your approach?

We run a semi-pseudo LoQ at the reporting limit of our impurities for pharma compounds. This is normally 0.05% concentration of our typical sample solution.

However, this isn't a true LoQ in the strictest sense of the definition. All it means is that we have a S/N greater 10 so it gives confidence that we can adequately quantify our impurities at this level. However the true LoQ (S/N = 10) value is usually much lower.


Plus it will also depend on how you measure your S/N. S/N calculated as root mean square (RMS noise) >> than S/N calculated as EP noise. I just wish the regulators would get together and make sense of all these differences.
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry