FLow rate problem

Posted: Fri Jul 11, 2008 6:59 pm
by rice
My HLPC won;t seem to work at higher flow rates, It will not even purge/prime the lines as the flow rate is to high, it'll get to a point in pressure and shut off, when it is set to except these pressures. Any ideas?

Posted: Fri Jul 11, 2008 7:52 pm
by zokitano
Please detail your question!
What brand of HPLC system are you using? Is column installed? At which pressures you have observed pump shut down, etc

Posted: Sat Jul 12, 2008 2:04 pm
by unmgvar
on some pumps there is a frit that when blocked produces a lot of pressure with a column or at high flow rates when purging/primimg.

it could be that a certain part of the tubing is blocked also.

or simply maybe the settings of maximum pressure for your pump are too low and it shuts itself because of that.

what is your pump brand?