FLow rate problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
My HLPC won;t seem to work at higher flow rates, It will not even purge/prime the lines as the flow rate is to high, it'll get to a point in pressure and shut off, when it is set to except these pressures. Any ideas?

Please detail your question!
What brand of HPLC system are you using? Is column installed? At which pressures you have observed pump shut down, etc

on some pumps there is a frit that when blocked produces a lot of pressure with a column or at high flow rates when purging/primimg.

it could be that a certain part of the tubing is blocked also.

or simply maybe the settings of maximum pressure for your pump are too low and it shuts itself because of that.

what is your pump brand?
3 posts Page 1 of 1

Who is online

In total there are 26 users online :: 4 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Majestic-12 [Bot], Semrush [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry