is my MSD have a problem??
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
7 posts
Page 1 of 1
-
- Posts: 1774
- Joined: Fri Feb 09, 2007 10:33 pm
che313,
You do not have enough sensitivity. Your 69 is too low so your 502 is just noise. Re-run an autotune and if your 69 does not come way up (along with your 502) then you need to start looking at some things. Your source might be dirty or your EM might be going bad. Autotune will greatly help you figure out the next step.
Best regards.
You do not have enough sensitivity. Your 69 is too low so your 502 is just noise. Re-run an autotune and if your 69 does not come way up (along with your 502) then you need to start looking at some things. Your source might be dirty or your EM might be going bad. Autotune will greatly help you figure out the next step.
Best regards.
-
- Posts: 136
- Joined: Mon Nov 19, 2007 3:16 pm
Thank you AICMM
I retuned it again this morning and EMVolts is 1882, H2O 1.96%, N2 7.49%, O2 0.24%, CO 0.1%, N2/H2O382.95%. and many others information if u need it I can post it.
I retuned it again this morning and EMVolts is 1882, H2O 1.96%, N2 7.49%, O2 0.24%, CO 0.1%, N2/H2O382.95%. and many others information if u need it I can post it.
-
- Posts: 1774
- Joined: Fri Feb 09, 2007 10:33 pm
che313,
Run a standard spectrum autotune and post that so we can have a look see.
Best regards.
Run a standard spectrum autotune and post that so we can have a look see.
Best regards.
-
- Posts: 136
- Joined: Mon Nov 19, 2007 3:16 pm
-
- Posts: 6
- Joined: Thu May 01, 2008 4:28 pm
che313,
Your m/z 69 mass assignment is good but the abundance is really low. Are the peak heights of your samples about the same as they had been in the past? If your sample peaks are the same, I would think that you may have a stuck calibration valve or are running low on PFTBA.
George Perham
Your m/z 69 mass assignment is good but the abundance is really low. Are the peak heights of your samples about the same as they had been in the past? If your sample peaks are the same, I would think that you may have a stuck calibration valve or are running low on PFTBA.
George Perham
-
- Posts: 176
- Joined: Mon Mar 14, 2005 7:01 pm
Hi!
In edit tune parameters menu, open manually the PFTBA valve and click on profile button.
Check intensity.....normally, intensity will be ok.
I suspect that your pftba valve is not openenig when you are doing autotune.
What is the version of your msdchemstation?
In edit tune parameters menu, open manually the PFTBA valve and click on profile button.
Check intensity.....normally, intensity will be ok.
I suspect that your pftba valve is not openenig when you are doing autotune.
What is the version of your msdchemstation?
7 posts
Page 1 of 1
Who is online
In total there are 210 users online :: 0 registered, 0 hidden and 210 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 210 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 210 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science