Dimethylformamide Carryover

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Dear all
iam not able to get rid of DMF carry over from Headspace (model : Agilent 7694), please suggest some temedy for this

Usually time and flow will cure carryover.

Is your hardware clean?

Can you give specifics?

best wishes,

Rod

Hi

Would like to add that you could check the sample valve temperature in the headspace. As DMF has a boiling point of ~153°C it could contribute. Especially if you have run samples before with DMF as the sample solvent, seen alot of problems related too low sample valv temperatures when high boilers DMF, DMSO, DMAA is used as sample solvent.

As the previous speaker said, hard to say without details.

Cheers Chris

Have you tried the steam cleaning procedure detailed in the manual? This usually clears most crap out of the system.
4 posts Page 1 of 1

Who is online

In total there are 190 users online :: 0 registered, 0 hidden and 190 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 190 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry