Troubles with Gradient Qualification on Dionex U3000
Posted: Tue Jun 17, 2008 8:07 pm
I have had some unusual problems associated with a gradient qualification that I was attempting to run on a Dionex U3000. The method I was attempting to run was laid out in a LC/GC supplement entitled "The Application Notebook." The method is described in the section "Performance Qualification of HPLC instrumentation in Regulated Laboratories" and appears on page 70 of the supplement, there is also a link to the article below.
The article suggests injecting a uracil spiked sample and running one of the mobile phases (in my case B) with a uracil spike. A gradient from 0 to 100% B is then started over ten minutes and held at 100% B for 5 minutes before going back to 10% B at 15 minutes and holding for 3 minutes. A set of plateaus with 10 to 90% B is then set up with intervals of 3 minutes. The article uses a column, but I used a restrictor coil in its place, I monitored at 273nm. I had water for mobile phases A and C and I had B as a solution of 0.035mg/mL Uracil in Acetonitrile. The article had a sample chromatogram, which I also have attached below.
My concern is the fact that my chromatogram (also attached below) has a few features that give me cause for concern. The first is the drop-off in absorbance once the % B reaches 100. There is then a spike in absorbance at 15 minutes when the %B changes from 100 to 10%, and similar but smaller spikes with the %B changing from 90 to 10%. I would welcome any suggestions on why the absorbances are behaving this way.
http://chromatographyonline.findpharma. ... &sk=&date=
Sample Chromatograph of the Gradient Qualification

My Chromatograph of the Gradient Qualification

The article suggests injecting a uracil spiked sample and running one of the mobile phases (in my case B) with a uracil spike. A gradient from 0 to 100% B is then started over ten minutes and held at 100% B for 5 minutes before going back to 10% B at 15 minutes and holding for 3 minutes. A set of plateaus with 10 to 90% B is then set up with intervals of 3 minutes. The article uses a column, but I used a restrictor coil in its place, I monitored at 273nm. I had water for mobile phases A and C and I had B as a solution of 0.035mg/mL Uracil in Acetonitrile. The article had a sample chromatogram, which I also have attached below.
My concern is the fact that my chromatogram (also attached below) has a few features that give me cause for concern. The first is the drop-off in absorbance once the % B reaches 100. There is then a spike in absorbance at 15 minutes when the %B changes from 100 to 10%, and similar but smaller spikes with the %B changing from 90 to 10%. I would welcome any suggestions on why the absorbances are behaving this way.
http://chromatographyonline.findpharma. ... &sk=&date=
Sample Chromatograph of the Gradient Qualification

My Chromatograph of the Gradient Qualification