- Posts: 3
- Joined: Tue Jun 17, 2008 6:48 pm
The article suggests injecting a uracil spiked sample and running one of the mobile phases (in my case B) with a uracil spike. A gradient from 0 to 100% B is then started over ten minutes and held at 100% B for 5 minutes before going back to 10% B at 15 minutes and holding for 3 minutes. A set of plateaus with 10 to 90% B is then set up with intervals of 3 minutes. The article uses a column, but I used a restrictor coil in its place, I monitored at 273nm. I had water for mobile phases A and C and I had B as a solution of 0.035mg/mL Uracil in Acetonitrile. The article had a sample chromatogram, which I also have attached below.
My concern is the fact that my chromatogram (also attached below) has a few features that give me cause for concern. The first is the drop-off in absorbance once the % B reaches 100. There is then a spike in absorbance at 15 minutes when the %B changes from 100 to 10%, and similar but smaller spikes with the %B changing from 90 to 10%. I would welcome any suggestions on why the absorbances are behaving this way.
http://chromatographyonline.findpharma. ... &sk=&date=
Sample Chromatograph of the Gradient Qualification

My Chromatograph of the Gradient Qualification