I am using PDA detector. My base line is always going negative so I am not able to see my peak..what is causing that? it is the flow cell or something else.
-
- Posts: 405
- Joined: Wed Dec 05, 2007 4:55 pm
What kind of PDA do you have? What wavelength are you measuring?
Is this an isocratic or gradient separation? What are the mobile phase components?
In some situations, a negative drifting baseline is normal. In other cases, it could mean that one solvent ("A" if this is a gradient) is contaminated.
Is this an isocratic or gradient separation? What are the mobile phase components?
In some situations, a negative drifting baseline is normal. In other cases, it could mean that one solvent ("A" if this is a gradient) is contaminated.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
ACCTA, Inc.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Why do your peaks disappear? Do you use a strip chart recorder for data collection?
-
- Posts: 292
- Joined: Wed Jan 19, 2005 2:20 pm
MK - can you supply a example chrom?
See my earlier postings about strange baseline effects with a faulty flow cell on my UPLC system.
See my earlier postings about strange baseline effects with a faulty flow cell on my UPLC system.
-
- Posts: 104
- Joined: Sat Apr 28, 2007 12:49 am
I am using Accela PDA(by thermofisher). My mobile phase is 70:30 Water:MeOH (isocratic).Flow rate is 0.2ml/min. Base line starts at 0 and then goes down to -2000-5000 and then I have normalise( auto intensity) to see the peaks.
-
- Posts: 104
- Joined: Sat Apr 28, 2007 12:49 am
wave lenght I am using 220 and 254.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
What I tried to ask is whether your data are lost when you don´t see the peaks while the chromatogra is running. I am just curious about what people hook up to a relatively modern machine like a PDA to do data analysis/workup.
-
- Posts: 835
- Joined: Thu Apr 14, 2005 7:00 am
-2000 to -5000 mAU is a lot.
accela models are pretty new so your instrument should be still under warranty logically.
first approach that i would take is use the vendors PQ test for drift.
if the test fails and the procedure uses water then there are 2 reasons for the failure.
the PDA itself or the water. make sure that you use good quality HPLC grade water for the test if it is required in the procedure.
accela models are pretty new so your instrument should be still under warranty logically.
first approach that i would take is use the vendors PQ test for drift.
if the test fails and the procedure uses water then there are 2 reasons for the failure.
the PDA itself or the water. make sure that you use good quality HPLC grade water for the test if it is required in the procedure.
8 posts
Page 1 of 1
Who is online
In total there are 27 users online :: 3 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], John Guajardo, Semrush [Bot] and 24 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], John Guajardo, Semrush [Bot] and 24 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science