Anyone had any experience with HCCPD by EPA 508? Always is giving us a headache in DW PT studies. Getting like 150-172% recovery on PT studies, QC are good. Argh!

Are you using a second source, and does it confirm your calibration?

Yes it does. I know we had a problem with split/splitless injector on the pesticide method (508, etc). Might there be a similar problem here?

My mistake on that last post. We did discover a problem with split/splitless on one of the other instruments - not necessarily the one for 508. Sorry.

Do you see a quadratic curve over the range of your calibration? I have had experience with HCCPD being a poor responder in the past, but if it's coming up high, and your spikes are okay, I would look at the range of your plot.

I'll go back to the lab and pull some curves and check. I'll let you know - thanks
6 posts Page 1 of 1

Who is online

In total there are 83 users online :: 0 registered, 0 hidden and 83 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 83 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry