- Posts: 122
- Joined: Fri Sep 21, 2007 7:43 pm
RF = (A[C23:0] * M[X]) / (A[X] * M[C23:0])
A is the area response of C23:0 or FAME X, and M is the mass.
The RF is then used to quantitate the samples. For system suitability, the RF must fall in the range 0.95 to 1.05 (I have no control over this, this is what was given to me, and the transferring lab is able to achieve this). The FAMEs in question are close in molecular weight and response in the chromatogram to the internal standard. The transferring lab is also using Agilent 6890 instruments.
My sample analyses have been closely matching the assays for known samples, which is great, but I can't get the system suitability to pass, which is something I absolutely have to overcome. I am injecting a 21 FAME standard from Nu-Chek additionally as a check... I know the mass ratios of the two FAMEs of interest to the C23:0 (the C23:0 is itself a component of the standard from Nu-Chek, I didn't add it) and I also have to calculate RFs for the 21-FAME mix. These RFs are badly off, as much as 1.11 (always high). These RFs also have to fall between 0.95 and 1.05. As well, the area percents of the C23:0 and two analytes have to be close to the expected values and I am off by a full percent on two of them.
This is consistent between different vials and dilutions of the the test mix, so I don't think my problem is likely due to the sample prep, unfortunately. I am using Agilent crimp cap vials with teflon-lined crimp caps. I am seeing every peak in the 21-FAME mix and my blanks have no interfering peaks, so I believe that coelutions cannot be responsible for my problems and therefore fiddling with the flow or over program would be a pointless exercise - besides which, I have done extensive work with the flow and oven program specifically to eliminate coelutions and my problem is consistent.
It has been suggested to me that if I can't get the RFs to work for the sys suit then I may have an inlet activation problem. I am using a split/splitless inlet in split mode with a split ratio of 100:1 (this is non-negotiable). I am using an SGE FocusLiner (Flow arrow pointing down), these have restrictions around the glass wool plug and come individually foil wrapped, I was told by the experience party at another lab that these liners were the best choice they had encountered to significantly reducing these kinds of problems. I am replacing the septum and the liner before every run.
Assuming inlet activation is my problem, can anyone suggest any further steps I could take to fix it and get my sys suit to work... unfortunately I have almost no latitude with the method, the inlet temperature of 250 C is a given and my oven ramp and flow rate, which I do have some latitude with, can't be touched because the paramters are already optimal - any change and I get unacceptable coelutions (trust me, I know!!). I don't believe any of these parameters are likely to be pertinent in any case.
I am not highly experienced with the mechanics of GC - I don't have a full grasp on what inlet activation actually is, and beyond replacing the liner with a new liner from a sealed package, which I've already tried and failed, I don't have a lot of other ideas as to what to do. Thanks for any help anyone can give!!