Column Performance.

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello All;
I have a issue with my columns, we are trying to develop a standart procedure for criteria of when you have to regect a column or when you need to Regenerate it; I know thres is now a Standart procedure but I suposse that may be USP or FDA have a criteria bese on N; Taling, selectivity; that we can use to determine when we have to get a new column or do a regeneration.
Usually I develop the method and not concern about this; but now I have to help this this.

I will Appreciate you advices.

Thanks and Best Regards.

As far as I know, there is no "one size fits all" standard. If you are dealing with a validated, regulatory method, you should have system suitability criteria for that method. When you no longer pass system suitability, the column is no longer suitable :wink: .

The FDA "suggestions" for system suitability date back to 1987 and are, in my opinion, only there as examples (N > 2,000 !).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry