I am a new employee at a certain laboratory and I am evaluating three "used" Waters PDA (996) detectors, with nothing but past experience to campare to.
Using Waters Millennium software, we will consider the criteria for a LOQ (Limit of Quantitation) peak to be a signal-to-noise NLT 10:1.
This calculation is S/N = (peak height (uv))/(baseline noise (mv) *1000)
I understand that precision and accuracy studies at this LOQ level will confirm an allowable S/N across detectors, however...
Generally speaking, from your experience, is there a minimum peak area(uv/s) or minimum peak height (uv) that should be expected to give a S/N of NLT 10:1...
The reason I ask is that of the three Waters PDA (996) detectors I have, the same LOQ solution gives the following results:
Detector--------Baseline Noise------Peak Height---S/N---Peak Area (uv/s)
PDA 1-----------0.019------------------323-----------17------------~10000
PDA 2-----------0.032------------------323-----------10------------~10000
PDA 3-----------0.048------------------323-----------07------------~10000
I understand that in using only the S/N NLT 10:1 criteria, only detectors 1 and 2 would be "suitable" for use with this method. However, I believe this peak height of ~320 and the area of 10000 should give much larger S/N values than 7 - 17.
In other words, I believe these Waters PDA (996) detectors are "noisier" than what I have observed in the past at different laboratories using Waters PDA (996) detectors. Are these Waters PDA (996) detectors performing par for the course or are they noisier than what you've seen?
Thanks in advance for your thought!!