Agilent 1100 HPLC - Is it necessary to gradually increase pressure and flow rate when switching methods and before the first run after the system has been turned on? Why?
-
- Posts: 835
- Joined: Thu Apr 14, 2005 7:00 am
for fragile type of columns it is advisable to increase or decrease the flow rate thus pressure in a gentile manner.
the column might be "fragile" for several reasons, the casing or the packing for example.
in the "old days" the procedure of slowly increasing/decreaing flow was extremely common for all types of columns, including all the silica based, because the packing was not as good as today. then the pressure differences produced by increasing or decreasing the flow would cause the packed bed to compress or relax creating dwell volumes and ruinning columns.
these days we mainly watch the maximum pressure allowed.
if you are uncertain, the best aproach i think is to ask the vendor.
the column might be "fragile" for several reasons, the casing or the packing for example.
in the "old days" the procedure of slowly increasing/decreaing flow was extremely common for all types of columns, including all the silica based, because the packing was not as good as today. then the pressure differences produced by increasing or decreasing the flow would cause the packed bed to compress or relax creating dwell volumes and ruinning columns.
these days we mainly watch the maximum pressure allowed.
if you are uncertain, the best aproach i think is to ask the vendor.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
We don't do that. And that certainly defeats the automated switching between methods in the Sequence Table (and/or columns using the column switching valve accessory). I agree with first response: was more important in the "old days".Agilent 1100 HPLC - Is it necessary to gradually increase pressure and flow rate when switching methods and before the first run after the system has been turned on? Why?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Never done this in my entire life with silica-based columns.
It is advisable though with compressible packings, i.e. polymer-based columns.
It is advisable though with compressible packings, i.e. polymer-based columns.
-
- Posts: 2174
- Joined: Tue Aug 17, 2004 7:59 pm
I suspect you have, Uwe - Alliance pumps have a setting for Pump Acceleration (Empower) or Flow Ramp (2695 panel). It appears as if it is only for flushing and purging, but I think it is also applied when initiating flow with a column in line. I think most modern pumps take at least a couple of turns of their drive motors before they're operating at their programmed rates, if they are starting from a stand-still.
Thanks,
DR

DR

-
- Posts: 208
- Joined: Wed Jul 06, 2005 3:35 pm
Only with gel columns (low cross linked stationary phases like the ones used for carbohydrate and amino acid analysis).It is advisable though with compressible packings, i.e. polymer-based columns.
Not necessary for highly cross linked phases (>50%XL).
As DR pointed out, most (all?) modern pumps need a few moments to get up to full flow rate.
--
Robert Haefele
Robert Haefele
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
It´s a myth, then, that pumps damage packings?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Probably not so much a myth as an old tradition that has (mostly) outlived it's usefulness.It´s a myth, then, that pumps damage packings?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 45
- Joined: Thu Mar 03, 2005 11:11 pm
it depends on the method. I have a method that I always have to increase the flow rate graduately at the beginning on 1100 instruments to avoid over pressuring. then again I am pushing the 1100 close to its limits with this method (normal operation pressure around 320 -350 bar depending on the actual system configuration). that's the only way to achieve the desired separation with a 30 min gradient. if I go with the conventional 1 mL/min flow rate and 3 um column, it would take 50 min.
when this method was given to the chemists for them to run their reaction mixture samples, they complained it always shut down their instrument when they first started it up. I set up a flow ramp method and told them to use it at the first line of their sequence in ChemStation. the problem went away.
when this method was given to the chemists for them to run their reaction mixture samples, they complained it always shut down their instrument when they first started it up. I set up a flow ramp method and told them to use it at the first line of their sequence in ChemStation. the problem went away.
Regards,
JZT
JZT
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
It is interesting to see, though, that all columns take/took the injection pressure shock quite well. This pressure fluctuation can often be detected even at the outlet of the column via the system peak it produces, for instance with a UV detector.
jzt, is the mobile phase at an optimum condition or was the high flow rate a "convenience" solution?
jzt, is the mobile phase at an optimum condition or was the high flow rate a "convenience" solution?
-
- Posts: 405
- Joined: Wed Dec 05, 2007 4:55 pm
Remember that most modern silica-based columns are packed at much higher pressures compared to the usual operating limits, so an initial "shock" to the column is probably not an issue. Most users do not have to worry about this problem.
The pressure changes that jzt sees at the beginning are probably due to differences in the viscosity of mixtures of the solvents. In this case, a slower flow would be advisable.
The Agilent 1100 pump has a "soft start" design, so it has a flow ramp at startup. Most modern pumps have this feature.
The pressure changes that jzt sees at the beginning are probably due to differences in the viscosity of mixtures of the solvents. In this case, a slower flow would be advisable.
The Agilent 1100 pump has a "soft start" design, so it has a flow ramp at startup. Most modern pumps have this feature.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
ACCTA, Inc.
11 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science