I am currently developing a method for the analysis of glycols (propylene glycol up to and including triethylene glycol) in aqueous samples.
Initally I went for split/splitless injection since we had an instrument available. After numerous problems and some research I've switched to on-column with some better and some poorer results.
The current set-up is cold on-column analysis with the inlet at 80 Deg C injecting onto a 2.5m polar deactivated pre-column which is attached to a 25m wax column. The start temperature for the oven is also 80 Deg C. Peak shape and calibration curves are good for the first five compounds but diethylene glycol and triethylene glycol both have very small peaks relative to the early eluters. The only way I have been able to improve this is by raising the temperature of the inlet to >200 Deg C. Other experiments have shown that the problem isn't with the syringe and the degree of carryover/ghost peaks confirms this. There is next to no ghosting of the first analyte and the degree of ghosting increases as the molecular weight increases.
My question is this:- by my understanding the sample should flow rapidly through the uncoated precolumn to 'trap' on the analytical column but the evidence seems to suggest that the heavier analytes are sitting on the pre-column in the inlet (which remains at 80 Deg C throughout the run) and being rinsed onto the analytical column when another injection is made.
Is my understanding of this concept flawed ?
Any comments appreciated.
Rich