I am working on a method for a Phospholipid that is ~95% Phosphatidylcholine (PC) and ~4.0% Lysophosphatidylcholine (LPC).
The method I am using was obtained from the vendor, and is a normal phase method that uses ELSD. However, after following the method carefully, my chromatagrams are coming out horrible. With ELSD, I understand reproduceability goes down, but is there anything I can do to get a somewhat normal baseline?
Second, I also tried cCAD with the same conditions as I did for ELSD, but my baseline starts off high, and i end up getting what almost looks like an Infrared spectrum.
In all of my chormatagrams, I also have a huge drop in signal at about 20 min which I think may be because of the gradient.
Any suggestions on how to improve this chromatography for a relatively new chemist? Method conditions are listed below:
-------------------------------------
Column: Phenomenex Luna HILIC 200A 5.0um 150 x 3.0mm with HILIC 4.0 x 2.0mm guard cartridge
Column Temp: 55 deg C
Flow Rate: 1.0 mL/min
Injection: 20.0 uL
ELSD Settings:
Temp: 50 deg C
Gain: 1
Gas Flow: 2.0 L/min
Mobile Phase A:
n-hexane 81.42%
2-propanol 17.00%
acetic acid 1.50%
Triethylamine 0.08%
Mobile Phase B:
2-propanol 84.42%
purified water 14.00%
acetic acid 1.50%
triethylamine 0.08%
Gradient (%B)
0.0min (5%)
5.0min (20%)
8.5min (40%)
15.0min (100%)
17.5min (100%)
17.6min (5%)
21.0min (5%)
22.0min (5%)
27.0min (5%)
29.0min (5%)
Sample Diluent:
50:50 n-hexane/2-propanol
Sample concentration is 2.1 mg/mL
Thanks in advance.