Agilent 5973MSD Air Leak
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
9 posts
Page 1 of 1
I'm having a difficult time troubleshooting a signiifcant air leak on 5973MSD. Noticed significant N2 and O2 in leak check diagonostics. Needed to change columns any way, so vented. Installed MSD inteface nut with no-hole ferrule only and pumped down. Leak check good (<2% 28(N2 abundance). Vented again. Installed column to MSD inteface nut with inlet end of column plugged to isolate it from potential inlet leaks. Pumped down but get VERY high 28 (N2). Sprayed dichlorodifluoromethane on MSD interface fitting and monitored m/z 85. but none detected. Where could air leak be coming from?
-
- Posts: 374
- Joined: Fri Oct 22, 2004 7:14 pm
Is is possible the column was broken?
-
- Posts: 25
- Joined: Tue Feb 27, 2007 8:53 pm
No. I checked that. Also tried same this morning with another column I had just taken out of another GC-MSD that was working. Same symptoms.Is is possible the column was broken?
-
- Posts: 1774
- Joined: Fri Feb 09, 2007 10:33 pm
djletin,
Very high 28 AND 32, right? Possible that the MSD nut is stripped, possible that the ferrule is installed the wrong way? Possible that the ferrule is cracked? How did you plug the inlet end of the column?
Just some thoughts.
Best regards.
Very high 28 AND 32, right? Possible that the MSD nut is stripped, possible that the ferrule is installed the wrong way? Possible that the ferrule is cracked? How did you plug the inlet end of the column?
Just some thoughts.
Best regards.
-
- Posts: 25
- Joined: Tue Feb 27, 2007 8:53 pm
Plugged inlet end of column with septum. Was thinking either MSD nut stripped or perhaps theads on transfer line damaged. Will try to repeat making sure I use brand new MSD nut.djletin,
Very high 28 AND 32, right? Possible that the MSD nut is stripped, possible that the ferrule is installed the wrong way? Possible that the ferrule is cracked? How did you plug the inlet end of the column?
Just some thoughts.
Best regards.
-
- Posts: 92
- Joined: Sat Mar 01, 2008 12:31 pm
I have used a "die" to correct cross threading on the interface. Also toss out the old interfacen nuts since they may be misaligned.
-
- Posts: 3596
- Joined: Mon Aug 30, 2004 11:43 pm
I'm sure you remembered that the ferrule at the MSD interface is positioned "backwards", correct?
-
- Posts: 176
- Joined: Mon Mar 14, 2005 7:01 pm
I suspect gas tank contamination....change the gas tank, wait 30min for clear old gas and check again.
-
- Posts: 5
- Joined: Wed Aug 10, 2005 12:16 pm
Possible reasons for poor resolution:
1. contaminated mass spectrometer,
2. poor mean free path i.e. poor vacuum,
3. wrong tune files i.e. the tune parameters are not matching with current instrument parameters.
Isolate the leak as nitrogen or moisture and look out for the nitrogen:oxygen ratio, in general 3:1 or 5:1 indicative of air leak. If nitrogen dominates suspect contaminated gas filters.
If isolated MSD had leak, look out for damaged O rings at vent, side plate, vacuum gauge manifold, CI connection.
If GC had problem, look out for aged septa, leak at liner nut, injector nut, detector nut and contaminated gas filter.
The Agilent mass spectrometer calibrated using perfluorotributyl amine (PFTBA) using masses 69, 219 and 502 m/z utilizing Mathieu stability diagram.
Among the tunes two are important on the basis of sensitivity – autotune and resolution – standard spectra tune with difference in peak width of 0.6 and 0.5 respectively. In general, the auto tune itself the trading off between sensitivity and resolution.
Happy MSD days
1. contaminated mass spectrometer,
2. poor mean free path i.e. poor vacuum,
3. wrong tune files i.e. the tune parameters are not matching with current instrument parameters.
Isolate the leak as nitrogen or moisture and look out for the nitrogen:oxygen ratio, in general 3:1 or 5:1 indicative of air leak. If nitrogen dominates suspect contaminated gas filters.
If isolated MSD had leak, look out for damaged O rings at vent, side plate, vacuum gauge manifold, CI connection.
If GC had problem, look out for aged septa, leak at liner nut, injector nut, detector nut and contaminated gas filter.
The Agilent mass spectrometer calibrated using perfluorotributyl amine (PFTBA) using masses 69, 219 and 502 m/z utilizing Mathieu stability diagram.
Among the tunes two are important on the basis of sensitivity – autotune and resolution – standard spectra tune with difference in peak width of 0.6 and 0.5 respectively. In general, the auto tune itself the trading off between sensitivity and resolution.
Happy MSD days
Dr.N.Bhuvaneswaran
India
India
9 posts
Page 1 of 1
Who is online
In total there are 217 users online :: 1 registered, 0 hidden and 216 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Semrush [Bot] and 216 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Semrush [Bot] and 216 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science