I am doing HS-GC method development and validation of residual solvents, methanol (MeOH) and isopropanol (IPA), in an API.
The RSDs for peak area of the two analytes from 6 replicat injections of the standard solution are really bad. Usually they are greater than 5%.
Method parameters are listed below:
Diluent: Water
Standard Solution: MeOH 150 μg/ml, IPA 250 μg/ml
GC: 7890A, FID Detector
Inlet temperature: 200 ℃, Detector temperature: 250 ℃
Oven Temperature: 40℃ for 5 min, 38℃/min to 230℃, hold for 8 min
Colum: DB-624, 30m*530μm*3μm
Carrier gas: Nitrogen, Flow rate: 4.8 ml/min, Split ratio: 10 to 1
Hydrogen flow rate: 40 ml/min, Air flow rate: 400 ml/min
Make-up flow rate: 35.2, Septum purge flow: 5 ml/min
HS: G1888, Oven temperature: 80℃, Loop temperature: 90℃
Tran. Line temperature: 105℃, Shake: High, Vial pressure: 15.2 psi
Equilibration time: 25 min, Vial pressurization time: 0.4 min
Loop fill time: 0.2 min, Loop equilibariton time: 0.1 min
Injection time: 1.0 min, Injection volume: 1.0 ml
A point to note is that this HS-GC instrument gives me wonderful RSD <2%, when I am runing similar methods with other diluents such as DMF and DMI. So I am keen to find out any unsuitability of the diluent (water) for the method parameters above.
Hope I have said what I want to say due to my unpolished English. Please shed some light on this.
Thanks in advance.
Terry